I'm having some issues with a LS Plus, whenever I perform a head protocol scan, everything goes great with the image adquisition and saving. But most of the times when I try changing to another protocol, specially with the TORAX and ABDOMEN ones, the system aborts the scan after the very first X ray exposure.
I've already changed and tested the IG in another system and is working perfectly, and changed the DARC HDDs because they were failing.
I performed the following tests and they were successfully completed:
Recon Data Path (good)
DAS Tools (good)
DDC (good)
Scan Data Path (good)
Error: 245190
Host: ct99
File: scan_complete.cxx
Function: System Control: Scan Control
System Aborted Scout Scanning, No views, either offset or data, were collected/received.
Error: 230019006
Host: ct99
File: sdio_proxy.c
A non-zero errno was encountered during a remote system call. The system call may be failed to complete successfully or may have been interrupted. The following diagnostic information was returned.
errno:5 (Input/output error)
Error: 230000897
Host: darc
File: raw_file.cxx
Function: Data Acquisition: Scan Data Save
Write to scan data disk failed
Error: 230000030
Host: darc
File: dip_thread.cxx
Function: Data Acquisition: Scan Data Acquisition
Unable to write offset view data to the scan file.
Error: 230011075
Host: ct99
File: sdio_xfs_io_ops.c
Function: Data Acquisition: SBC Processing
Unable to open a scan file on the scan data disks. Read of scan file header failed in read()call
If anybody had this issue before or knows something else that I'm missing, please let me know.
what is the primary error displayed in the error log?
Try to reduce the KV for thorax and abdomen exam, and perform xray and see if it continue abording scans. If it does'nt aborded with low KV and mAs, your problem is in the Xray generator (HV Tank, Tube, ...).
Verify the HV oil in each tip of HV tank and Xray tube; add if necessary.
ThanKS guys for your thoughts... I had the SDDA disks bad, even the new ones that I have changed once before... that solved the issue, it wasn't saving the raw data well in the disks... so once the first exposures we're taken, the scan aborted inmediatly.
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
FranciscoG
LightSpeed is aborting the scans
November 26, 2015 05:38
Hi guys,
I'm having some issues with a LS Plus, whenever I perform a head protocol scan, everything goes great with the image adquisition and saving. But most of the times when I try changing to another protocol, specially with the TORAX and ABDOMEN ones, the system aborts the scan after the very first X ray exposure.
I've already changed and tested the IG in another system and is working perfectly, and changed the DARC HDDs because they were failing.
I performed the following tests and they were successfully completed:
Recon Data Path (good)
DAS Tools (good)
DDC (good)
Scan Data Path (good)
Now these are the errors that its giving me:
Error: 5
Host: darc
File: file_save_thread.cxx
OS error: 5--- Input/output error
Error: 245190
Host: ct99
File: scan_complete.cxx
Function: System Control: Scan Control
System Aborted Scout Scanning, No views, either offset or data, were collected/received.
Error: 230019006
Host: ct99
File: sdio_proxy.c
A non-zero errno was encountered during a remote system call. The system call may be failed to complete successfully or may have been interrupted. The following diagnostic information was returned.
errno:5 (Input/output error)
Error: 230000897
Host: darc
File: raw_file.cxx
Function: Data Acquisition: Scan Data Save
Write to scan data disk failed
Error: 230000030
Host: darc
File: dip_thread.cxx
Function: Data Acquisition: Scan Data Acquisition
Unable to write offset view data to the scan file.
Error: 230011075
Host: ct99
File: sdio_xfs_io_ops.c
Function: Data Acquisition: SBC Processing
Unable to open a scan file on the scan data disks. Read of scan file header failed in read()call
If anybody had this issue before or knows something else that I'm missing, please let me know.
Thanks everybody!
to rate and post a comment
Youbissi Pascal
re: LightSpeed is aborting the scans
November 28, 2015 07:18
what is the primary error displayed in the error log?
Try to reduce the KV for thorax and abdomen exam, and perform xray and see if it continue abording scans. If it does'nt aborded with low KV and mAs, your problem is in the Xray generator (HV Tank, Tube, ...).
Verify the HV oil in each tip of HV tank and Xray tube; add if necessary.
to rate and post a comment
imeng
re: LightSpeed is aborting the scans
November 28, 2015 07:20
Hi,
please clean the slipring.
stay well
imeng
to rate and post a comment
Francisco Garcia
re: LightSpeed is aborting the scans
November 29, 2015 10:23
ThanKS guys for your thoughts... I had the SDDA disks bad, even the new ones that I have changed once before... that solved the issue, it wasn't saving the raw data well in the disks... so once the first exposures we're taken, the scan aborted inmediatly.
Thanks to you all.
to rate and post a comment