Lightspeed LFC problem

June 03, 2015 09:45

Hi all you guys!
I made a Load from Cold to the Lightspeed Plus CT system, because a problem and replacement of the RIP board, MVME23xx. The SW revisions are the IRIX 6.5.16M OS and LightSpeed Plus SDAS 2.4.2_H2.4M5 Applications.
My problem is that the system does not go up to Applications, instead it shows up a window that says "Welcome to ct01, Hispeed CT/i" and ask for Login and Password. But I installed a LightSpeed plus!!!
The console is with Octane computer and Pegasus IG recon board.
I'll appreciate a lot your comments

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Lightspeed LFC problem

June 04, 2015 05:39

do you have the LFC procedure.
here is a list of software please confirm what you actually have.
2.2 Gather Software CD-ROMs
2.2.1 LS2002 M5 Software
Operating and applications software is contained on five (5) CD-ROMs. To load system software,
you must have the following CD-ROMs:
1.) IRIX Operating System CD-ROM Disk #1 (P/N 2356015)
2.) IRIX Operating System CD-ROM Disk #2 (P/N 2356016)
3.) IRIX Operating System CD-ROM Disk #3 (P/N 2356017)
4.) LS2002 M5 or M5.1 Applications Software CD-ROM Disk #1 (see P/Ns below)
5.) LS2002 M5 or M5.1 Applications Software CD-ROM Disk #2 (see P/Ns below)
6.) LS2002 M5 Patch A CD-ROM (P/N 2399867) or LS2002 M5.1 Patch A CD-ROM (P/N
5112619)
Note: LS2002 M5 Patch C.1 CD-ROM (P/N 5113667) is available and recommended for sites with M5 or
M5.1 software (refer to Support Central or GE Service on how to obtain this patch).
NOTICE The Applications Software CDs, are particular to the individual configurations. Be certain
that the Application CDs match the system onto which they are being loaded:
• LightSpeed16 (MDAS) M5.1 (supports SMM2400GEMS-01 Motorola MVME 2433 SBC
256MB SBC, P/N 2370825)
- CD-ROM Disk #1, P/N 5112617
- CD-ROM Disk #2, P/N 5112618
• LightSpeed16 (MDAS) M5
- CD-ROM Disk #1, P/N 2395170
- CD-ROM Disk #2, P/N 2395171
• LightSpeed Ultra/Plus (MDAS)
- CD-ROM Disk #1, P/N 2395162
- CD-ROM Disk #2, P/N 2395163
• LightSpeed Plus (SDAS)
- CD-ROM Disk #1, P/N 2395160
- CD-ROM Disk #2, P/N 2395161
IRIX OS disks, however, are identical for all CT scanner configurations.
2.2.2 LS2002 M4 Software
Operating and applications software is contained on five (5) CD-ROMs. To load system software,
you must have the following CD-ROMs:
1.) IRIX Operating System CD-ROM Disk #1 (P/N 2356015)
2.) IRIX Operating System CD-ROM Disk #2 (P/N 2356016)
3.) IRIX Operating System CD-ROM Disk #3 (P/N 2356017)
4.) LS2002 M4 Applications Software CD-ROM Disk #1 (see P/Ns below)
5.) LS2002 M4 Applications Software CD-ROM Disk #2 (see P/Ns below)
6.) LS2002 M4 Patch C CD-ROM (P/N 2366813) – for MDAS4 and MDAS8 Only
NOTICE The Applications Software CDs, are particular to the individual configurations. Be certain
that the Application CDs match the system onto which they are being loaded:
• LightSpeed16 (MDAS)
• LightSpeed Ultra/Plus (MDAS)
- CD-ROM Disk #1, P/N 2360608-2
- CD-ROM Disk #2, P/N 2360609-2
• LightSpeed Plus (SDAS)
- CD-ROM Disk #1, P/N 2360614-2
- CD-ROM Disk #2, P/N 2360615-2
IRIX OS disks, however, are identical for all CT scanner configurations.

Log inor Register

to rate and post a comment

(128)
(10)
(15)

Johnny Barreto

re: re: Lightspeed LFC problem

June 05, 2015 10:25

Hi Lenny, thank you for your repply.
The disks I have are:
OS Irix 6.5.16M disk #1 PN 2356015
OS Irix 6.5.16M disk #2 PN 2356016
OS Irix 6.5.16M disk #3 PN 2356017
Application SW 2.4.2_H2.4M5 disk #1 PN 2395160
Application SW 2.4.2_H2.4M5 disk #2 PN 2395161
Aplications LS2002 M5 SW patch A PN 2399867
Aplications GOC1 M5 SW patch G.6 PN 5142334 Rev 2

The works I made for this system is:
- Replacement of RIP board, PN 2197234-2
- LFC following procedures of the LS2002 SW installation manual PN 2348269-100 rev 10
- I got an error on pflash, so I made pflash and cnfg;m procedures manually.

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: Lightspeed LFC problem

June 05, 2015 03:25

please forgive me if this is to basic, you did verify your jumpers on the RIP board?
I missed a jumper and it caused the same issue

Log inor Register

to rate and post a comment

(128)
(10)
(15)

Johnny Barreto

re: re: re: re: Lightspeed LFC problem

June 06, 2015 04:22

The jumpers are at same position. How could I discard problems in the piggyboards?

Log inor Register

to rate and post a comment

mriserv

re: Lightspeed LFC problem

June 05, 2015 10:16

It looks like Applications CD ddint not load properly. It should set up the environment paramaters and could not ask you logon or password after a successful load.
Try loading the software from scratch using the proper documentation

Log inor Register

to rate and post a comment

(1)
(1)

frank watts

re: Lightspeed LFC problem

June 06, 2015 04:53

I have had this problem before. On one occasion it was the Octane hard drive and then on the other it was the Octane itself.

1 person liked this
Log inor Register

to rate and post a comment

(128)
(10)
(15)

Johnny Barreto

re: Lightspeed LFC problem

June 09, 2015 04:38

My MVME board gets its red light lit. During application software load, I got error saying "Detection of Motorola board failed", also have another error standing something like "Pegasus IG could not reset". And the experience of Frank about the Octane. So I suspect for the Octane computer, the Pegasus IG and the RIP board again. Crazy. Is there a chance to upgrade this console to an HP basis console?

Log inor Register

to rate and post a comment

(3)

kumar sujeet

re: re: Lightspeed LFC problem

September 15, 2015 08:09

Hello dear,
Facing the same problem "Motorola detection failure"
Would be very thankful for ur help

Log inor Register

to rate and post a comment

stats
DOTmed text ad

JOHNSON & JOHNSON MONARCH QUEST - For Sale $299k or Lease $6k Per Month, 48 Months

Lightly used system available direct from the end user customer. MONARCH QUEST Featuring our most advanced, data-driven navigation software with AI-Powered algorithms & advanced imaging integration with the GE HealthCare OEC 3D C-arm.

stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)