LightSpeed Plus Fil/mA problem
July 31, 2014 04:06
This is a GE LightSpeed Plus, SDAS, 3.x system. It doesn´t make exposures neither in service mode. It gives an error message that stands "FIL/MA fault latch 1 would not reset" and "FIL/MA fault latch 2 woud not reset", both into the FUNCTION: XRAY GENERATION, MA loop.
Does anyone have any idea what these messages mean?
to rate and post a comment
Awodem Alain
re: LightSpeed Plus Fil/mA problem
July 31, 2014 10:00
You need to send more errors log so that I can help.
Run diag on the following :
HV Cathode tank
Ma/KV board
Xray tube
Your problem will be between those parts.
Only error log will help defined the bad one.
Good luck
to rate and post a comment
Johnny Barreto
re: re: LightSpeed Plus Fil/mA problem
September 05, 2014 05:23
Replacing the MA Board the problem has gone
to rate and post a comment