Joseph Porsenna
Logic 500 problems .
July 29, 2009 08:21
I am working on a logic 500 . The system will boot up and and recognize a 5mhz probe.
on trying to go to PD , the screen will show
sys error- SHVST-TRG error. at times,the doppler mode will work. The scan head voltage is reading
on the HV board 220 volts instead of 130 volts.
After running diagnostics , i got
doppler test 1 failed,doppler test 2 failed,
color flow test 2 failed,power supply 2 thermal test failed . B mode noise floor test failed.
has anyboby ever met that problem before ????
thanks for a quick reply.
to rate and post a comment
Ametepe AGBENOU
Re: Logic 500 problems .
August 01, 2009 04:14
Change your connector assy
to rate and post a comment
Angel Morales
Re: Logic 500 problems .
March 11, 2011 09:40
you have to test the high voltage card there is the problem