Thank you Gilberto, do you think you can send me your phone number so I can call you, iif you can please send to tcamargo3@gmail.com.
Thanks in advance.
Hello, Tomas! Replace R101, R102, R103 on the 3 port probe connector board for 47 Ohms, 1/4 Watt resistors. Check the old ones first, they are probably tending to infinity. If is a DCONN (2-port probe board), I'm not sure, but must be R31 and R32. Hope this is helpful, God Bless. MAC.
good day Marcelo,
I have the same problem and I followed the indications that you mention, the resistors are good, the voltages of the source are within the range, I need help to solve these problems, but you can give me a tip, to solve.
Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net
Tomas Camargo
LOGIQ 400 ERROR
June 25, 2012 03:25
Hi everyone, I hope someone can help!
I have a logiq 400 system that is giving me the following error.
Warning(AV ABNORMAL. PCONT [M12VL1LG])any help will be appreciated.
Thanks in advance.
to rate and post a comment
Gilberto Rubio Gil
re: LOGIQ 400 ERROR
June 25, 2012 05:34
probably you have a problem with the HV supply.
where are you from? I am in Guadalajara, Mexico. Maybe i can help you.
to rate and post a comment
Tomas Camargo
re: re: LOGIQ 400 ERROR
June 25, 2012 09:50
Thank you Gilberto, do you think you can send me your phone number so I can call you, iif you can please send to tcamargo3@gmail.com.
Thanks in advance.
to rate and post a comment
sitwat kashif
re: LOGIQ 400 ERROR
June 25, 2012 07:26
if you need service manual of L-400 contact with me ON FOLLOWING EMAIL: sitwatmedical@gmail.com
to rate and post a comment
Tomas Camargo
re: re: LOGIQ 400 ERROR
June 25, 2012 09:50
Thanks men but I already have the munual.
to rate and post a comment
Josip Vujnovic
re: re: re: LOGIQ 400 ERROR
June 26, 2012 09:33
Solution: UL-SOLUTION-002510
Title: UL_L400_HIGH VOLTAGE ABNORMAL PCONT_HVPWOVER
Solution Description:
H V Abnormal PCONT HVPWOVER
Resolution:
Possibly the BMRC brd
BMRC makes REFCV Control Signal for the TRDR
the REFCV controls the HV voltage.
to rate and post a comment
Miguel Flores
re: LOGIQ 400 ERROR
June 26, 2012 09:41
POWER SUPPLY ASSY/PCONT ASSY/Power Control Assembly
to rate and post a comment
Marcelo Antonio Correa
re: LOGIQ 400 ERROR
June 26, 2012 09:41
Hello, Tomas! Replace R101, R102, R103 on the 3 port probe connector board for 47 Ohms, 1/4 Watt resistors. Check the old ones first, they are probably tending to infinity. If is a DCONN (2-port probe board), I'm not sure, but must be R31 and R32. Hope this is helpful, God Bless. MAC.
to rate and post a comment
kco
re: re: LOGIQ 400 ERROR
November 22, 2016 10:08
good day Marcelo,
I have the same problem and I followed the indications that you mention, the resistors are good, the voltages of the source are within the range, I need help to solve these problems, but you can give me a tip, to solve.
to rate and post a comment
Madhavesh Kulkarni
re: LOGIQ 400 ERROR
June 26, 2012 09:36
Dear Tomas
pls check +-5/12/15vdc outputs on power supply assy. one/more output must b missing. you can repair it too.
Regards
Madhavesh
to rate and post a comment
Vimal Prakash Tripathi
re: LOGIQ 400 ERROR
January 12, 2013 11:25
dear tomas,
did u got resolve the problem regardings.
to rate and post a comment