Petar Stankovic
LOGIQ 400 problem. Please help!
September 27, 2021 09:28
After replacement it does not start. It hangs on the second step and prints no error on the screen.
I connected my laptop and here's the log of the terminal screen:
*** VENUS SYSTEM RESET/START ***
*** BBRAM contents are corrupted ***
VTSH>
VTSH>*** VENUS SYSTEM READY ***
PWRD : rt_reqprdmsg error [1100a]
to rate and post a comment
Humberto RoCe
re: LOGIQ 400 problem. Please help!
September 29, 2021 10:31
Connect lap top and use hyperterminal
Type the following(watch out for big and small letters)
VTSH>Id MSTPRG/BRST.LO
VTSH>st
Follow the onscreen instructions to set the time, date, etc.
to rate and post a comment