logiq 400

December 17, 2012 06:34

some time my machine give me error when i power up the machine...everything is working but probe image is not showing and give
{ error (shvst-pc) error system error }
anybody knows about it then plz help me .....

1 person liked this
Log inor Register

to rate and post a comment

(97)
(8)

Holger Fritze

re: logiq 400

December 17, 2012 06:56

most defective is the 5V and 12V DC-module on HV-Board. Please change both and feel happy..
Holger

Log inor Register

to rate and post a comment

Marcelo Antonio Correa

re: logiq 400

December 17, 2012 05:56

Hi, Khan! Check the 1 KOhm resistors (R31/32, if I am not wrong - they're close to the probe connector on the front of the board) They might be tending to infinity (huge resistance). If you have a 3-probe port board, check them the same way (their identifications are different, but they are close to the respective ports, too). In both cases, replace them for 47 Ohm, 1/4 Watt resistor. It should work perfectly and you won't have the message and the lock up anymore. Replace all, even if they are inside the 1 KOhm range. Hope this helps, Good holidays, God Bless. MAC.

2 people liked this
Log inor Register

to rate and post a comment

(10)
(4)

Usman Khan

re: re: logiq 400

December 27, 2012 10:16

thanks for your help....you might be right....i have to change resistor or anything more....

Log inor Register

to rate and post a comment

Marcelo Antonio Correa

re: re: re: logiq 400

December 27, 2012 03:13

Hi, Khan! I believe the resistors' replacement would be enough, trust. This action will fix the failure. Thank you for the reply, great 2013, God Bless you, a lot! MAC.

Log inor Register

to rate and post a comment

(6)

ahmad kh

re: re: logiq 400

November 23, 2017 05:28

thanks ... it work

Log inor Register

to rate and post a comment

ROBERTO DIAZ BEDOLLA

re: logiq 400

December 17, 2012 09:35

remove all probes... connect just one on first port and check if error continues... turn of the us... now connect the same probe to the second port and check i think one of 3 ports is defective

Log inor Register

to rate and post a comment

(22)
(1)

HUKAM CHAND

re: logiq 400

December 18, 2012 11:06

HV Power supply is bad.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

BK and other Urological Systems available. BK Probe repair available!!!

Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net

DOTmed Certified stats
advertisement

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)