Hello Sir,
i have the problem with logiq 400md, the problem is m/c gives error " cooling fan error and HV abnormal.PCONT(SHVVLILG).Kindly guide to solve the problem.
Thanking you,
Specializing in GE, we keep your ICU monitors calibrated and patient-ready—fast, reliable service from expert biomed techs. Southeastern Biomedical: 888-310-7322.
Rajan Alagu
Logiq 400md problem
August 21, 2008 09:44
Hello Sir,
i have the problem with logiq 400md, the problem is m/c gives error " cooling fan error and HV abnormal.PCONT(SHVVLILG).Kindly guide to solve the problem.
Thanking you,
to rate and post a comment
Lakshmi Narayana
Re: Logiq 400md problem
August 22, 2008 01:49
Hig Voltage power supply is faulty. I can try to repair this if you wish.
to rate and post a comment
Rajan Alagu
Re: Logiq 400md problem
September 13, 2008 07:02
Hello sir,
Thank you for guidance, i have replaced the 5v,+12v,-12v block,now m/c working well.
Thanking you sir,