Hello Sir,
i have the problem with logiq 400md, the problem is m/c gives error " cooling fan error and HV abnormal.PCONT(SHVVLILG).Kindly guide to solve the problem.
Thanking you,

Log inor Register

to rate and post a comment

(76)
(4)
(34)

Lakshmi Narayana

Re: Logiq 400md problem

August 22, 2008 01:49

Hig Voltage power supply is faulty. I can try to repair this if you wish.

Log inor Register

to rate and post a comment

(20)

Rajan Alagu

Re: Logiq 400md problem

September 13, 2008 07:02

Hello sir,
Thank you for guidance, i have replaced the 5v,+12v,-12v block,now m/c working well.