Hi, Leo! Have you tested the main monitor (plug an external VGA source as a Laptop to see what it happens). Did you mean that the US image is being displayed at 5 inches monitor? If so, the touch screen is the second monitor (auxiliary) and it's being considered as primary. Check all this first prior to suspect VIC board. Help this helps, take care, God Bless. MAC.
Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net
Leo Samba
Logiq 5
October 15, 2012 09:10
I have GE Logiq 5 with no patient exam displays on the monitor but showing on touch screen LCD. Is it the VIC card defective?
to rate and post a comment
Marcelo Antonio Correa
re: Logiq 5
October 15, 2012 08:36
Hi, Leo! Have you tested the main monitor (plug an external VGA source as a Laptop to see what it happens). Did you mean that the US image is being displayed at 5 inches monitor? If so, the touch screen is the second monitor (auxiliary) and it's being considered as primary. Check all this first prior to suspect VIC board. Help this helps, take care, God Bless. MAC.
to rate and post a comment
zhang hongjun
re: Logiq 5
March 05, 2013 08:39
Reload software or check VIC
to rate and post a comment