LOGIQ 7 REV7 Problem

Tags: Ultrasound

(28)
(1)

Luciano Alviggi

LOGIQ 7 REV7 Problem

April 19, 2015 09:50

My logiq7 starting good whitout error message on screen. But no see any probe. On touchscreen not appared probe name but only image of connector free. If I connect probe screen mess "probe disconnect" . And I no lissen the beep sound when finish boot. I reinstalled sw . Check all voltage . All ok . When I perform test on pc diagnostic its all passed. When I perform LOGIQ7 test I see "system configuration: fail
Help me please

Log inor Register

to rate and post a comment

(195)
(47)
(4)

EROTOKRITOS SKENTOS

re: LOGIQ 7 REV7 Problem

April 20, 2015 10:09

Have you changed anythink in the Logiq 7.The message you get is from an incopatible board or software wrong version or lost bios settings.

Log inor Register

to rate and post a comment

(28)
(1)

Luciano Alviggi

re: re: LOGIQ 7 REV7 Problem

April 20, 2015 12:08

I no chenge noting. I power on and have problem. No probe see on lcd

Log inor Register

to rate and post a comment

(28)
(1)

Luciano Alviggi

re: re: LOGIQ 7 REV7 Problem

April 20, 2015 12:08

I see on Utility-General-Info on-About info : QCON/PSEL : ff and on rev : ???

Log inor Register

to rate and post a comment

(195)
(47)
(4)

EROTOKRITOS SKENTOS

re: re: re: LOGIQ 7 REV7 Problem

April 21, 2015 07:29

The PC is communicating with the boards because you said you see empty the probe slots.If there was a problem with the communication you would see some probes that is not really connected and a black and white image,so the problem is in the boards.Remove the probe connector and reinstall it after cleaning it very well and try again.Also try to reinstall the software.

Log inor Register

to rate and post a comment

(28)
(1)

Luciano Alviggi

re: re: re: re: LOGIQ 7 REV7 Problem

April 21, 2015 11:54

I remove it clean and reinstall sw. Noting. If I run QCON test on screen i see : check rev . On Hardware info I see : QCON - ff - ???

Log inor Register

to rate and post a comment

(195)
(47)
(4)

EROTOKRITOS SKENTOS

re: re: re: re: re: LOGIQ 7 REV7 Problem

April 21, 2015 04:25

It seems like you found the problem.The message is clear QCON is the board that the probes connect.Do you have another one?

Log inor Register

to rate and post a comment

(28)
(1)

Luciano Alviggi

re: re: re: re: re: re: LOGIQ 7 REV7 Problem

April 22, 2015 07:13

I dont have

Log inor Register

to rate and post a comment

(28)
(1)

Luciano Alviggi

re: re: re: re: re: re: LOGIQ 7 REV7 Problem

April 22, 2015 04:16

Into Service - Log - Sistem , the first error log is " Probe Name is Empty "

Log inor Register

to rate and post a comment

Ricardo A. Sanchez

re: LOGIQ 7 REV7 Problem

April 21, 2015 07:29

is the BEP communicating with the FEC? are all the voltages present on the LV FEC power supply?

Log inor Register

to rate and post a comment

(28)
(1)

Luciano Alviggi

re: re: LOGIQ 7 REV7 Problem

April 21, 2015 11:55

Yes. BEP have comunication. Becouse in Hardware info I see all board connected. Only QCON have ??? in rev position. And on test write Check rev! . Supply is ok

Log inor Register

to rate and post a comment

(28)
(1)

Luciano Alviggi

re: re: LOGIQ 7 REV7 Problem

April 21, 2015 11:55

Yes. BEP have comunication. Becouse in Hardware info I see all board connected. Only QCON have ??? in rev position. And on test write Check rev! . Supply is ok

Log inor Register

to rate and post a comment