Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

Logiq 700 software problem.

Tags: Ultrasound, GE

(192)
(45)
(1)

Dante Nuevo

Logiq 700 software problem.

September 23, 2016 09:40

Good day, my dear dotmed friends.

I am here to ask for your help to confirm an error.
I have a set of disk to reload a logiq 700 ultrasound, but the first disk is not working and has a very strange list of files.
aaaaaaa0aa.iq
aaaaaaa0ab.iq
and goes like that 19 times to complete 26MB.
I am not sure if this MO disk is good since I didn't check it before and because it appears to start reading the disk and then stops.
The other 3 disks appers to be on the same condition. I tested the disks with 3 different drives with the same result.
Please tell me if you have seen it before and if it is good.

Thanks for your kind help.

Dante Nuevo.
dantenuevo@hotmail.com

Log inor Register

to rate and post a comment

(7)
(3)

Michele Di Lello

re: Logiq 700 software problem.

September 24, 2016 04:11

I do not know the file type, verify the operation of the floppy drive, this usually does not work because not used, it happened to me several times. regards

Log inor Register

to rate and post a comment

(119)
(57)
(6)

Jim Carr

re: Logiq 700 software problem.

September 24, 2016 04:11

Hello Dante,

What version of software are you trying to load? Before rev 6 the probes should not be connected when trying to install software.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

We provide sales, service & repair for portable ultrasound systems & probes

If you have a broken Sonosite, Mindray, GE, Philips, Hitachi, or Aloka system or probe contact us today. Free ultrasound & probe EVALS with 1 Year Warranty on all repairs. We repair all orders in 1-5 days. Call 844-644-5440 Email: Sales@Lenard.Healthcare

DOTmed Certified stats
advertisement

Upcoming Auction Events

Cbrx2sales
TIMED
Jun 05, 2026 (157 lots)