I have a problem with the Logiq Book XP.When I power it ON,the fans go on but nothing comes on the screen.I connected an external display but nothing comes on that too.Any suggestions?
Check the battery. The unit will not boot up or run normally with a low battery. If not battery, see if it will recognize the external CD used to reinstall the software.
Yes, that should be OK to not do that utility as it sounds like the application software got corrupted or you lost a sector on the hard drive that was fixed by the software reload.
There may be corrupted presets; try restoring the presets. There may be too many patient studies on the hard drive. Move them to a CD and ensure all patient studies are deleted. There is a database reset utility in the diagnostics (Utilities > Scanner Utilities) that will build a new database and will get rid of corrupted files that can cause the system to run slow or lockup. It that does not fix it, then try reloading software after making sure you have backed up the critical data.
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Ramya Penmesta
Logiq Book XP problem
June 04, 2010 02:04
Hello,
I have a problem with the Logiq Book XP.When I power it ON,the fans go on but nothing comes on the screen.I connected an external display but nothing comes on that too.Any suggestions?
to rate and post a comment
Thomas Vella
Re: Logiq Book XP problem
June 04, 2010 05:27
try reloading software
tom vella
hydacomp inc
to rate and post a comment
Jim Carr
Re: Logiq Book XP problem
June 04, 2010 10:13
Check the battery. The unit will not boot up or run normally with a low battery. If not battery, see if it will recognize the external CD used to reinstall the software.
to rate and post a comment
Ramya Penmesta
Re: Logiq Book XP problem
June 07, 2010 05:15
Thank you so much!!!Reloading software worked.But it dint allow me to load FPGA.Is that ok?
to rate and post a comment
Jim Carr
Re: Logiq Book XP problem
June 09, 2010 01:13
Yes, that should be OK to not do that utility as it sounds like the application software got corrupted or you lost a sector on the hard drive that was fixed by the software reload.
to rate and post a comment
Ramya Penmesta
Re: Logiq Book XP problem
June 09, 2010 10:19
Thank you so much.That is helpful!!
to rate and post a comment
RMSERVICE
re: Re: Logiq Book XP problem
July 29, 2013 03:32
hi jim, i have a problem with a logiq book xp, the system its very slow and 20 min to power on, how its the problem, help me please.
to rate and post a comment
Zach Ruder
re: re: Re: Logiq Book XP problem
July 29, 2013 05:50
Sounds like Jim's first response is probably going to be your best bet. First reload the software, and see what happens from there.
to rate and post a comment
Jim Carr
re: re: Re: Logiq Book XP problem
July 30, 2013 02:09
There may be corrupted presets; try restoring the presets. There may be too many patient studies on the hard drive. Move them to a CD and ensure all patient studies are deleted. There is a database reset utility in the diagnostics (Utilities > Scanner Utilities) that will build a new database and will get rid of corrupted files that can cause the system to run slow or lockup. It that does not fix it, then try reloading software after making sure you have backed up the critical data.