Ampronix, a Top Master Distributor for Sony Medical, provides Sales, Service & Exchanges for Sony Surgical Displays, Printers, & More. Rely on Us for Expert Support Tailored to Your Needs. Email info@ampronix.com or Call 949-273-8000 for Premier Pricing.
wen chen
logiq e LCD fault
April 22, 2014 10:05
i have a problem about LOGIQ E unite,the fault is the LCD screen shows pink occasinally,and conect a external LCD display shows normally,then i reset the LCD cable,but the fault is ramian,i suapact LCDcable. someone can give me some advice ,thanx
to rate and post a comment
Marcelo Antonio Correa
re: logiq e LCD fault
April 23, 2014 09:13
Hi, Chen!
I believe, due to the cover movement (opening and closing the screen), cable may be cutting and, sometimes, you see the pink image you described.
It's clear that you will need to dismount the LCD and check cable and connections and, probably, replacing some part involved with the VGA signal.
It might be, also, cold soldering in the VGA circuit, mainly noticed, if the pink appearance disappears after equipment is host.
Hope this helps, take care, God Bless.
Kind regards,
MAC.
to rate and post a comment
grazza
re: logiq e LCD fault
April 23, 2014 09:17
Yes, the LCD cable is very likely to be the problem. I have replaced them before for similar issues.