I have a logiq S8 that is having intermittent artifacts all over the scanning screen. Ran diagnostics for the noise floor and BF in low, med and high. Failed the noise test, channels 191 failed on low, med and high. I replaced the BF board(it wasn't cheap), a month later the issue comes back, it didn't fail the noise floor test but it did fail the same channels as before, plus channels 195 and 196 on the BF med test. Replaced the PID board and now we are seeing the same issue, has anyone seen this before?
Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net
seth solis
Logiq S8 artifacts
February 16, 2021 09:59
I have a logiq S8 that is having intermittent artifacts all over the scanning screen. Ran diagnostics for the noise floor and BF in low, med and high. Failed the noise test, channels 191 failed on low, med and high. I replaced the BF board(it wasn't cheap), a month later the issue comes back, it didn't fail the noise floor test but it did fail the same channels as before, plus channels 195 and 196 on the BF med test. Replaced the PID board and now we are seeing the same issue, has anyone seen this before?
to rate and post a comment