The 5VDC should be set for 5.2vdc. On the back of the machine is the LVPS (upper left corner). You will see a little access hole in the cover to adjust the pot. The other possible cause could be the Dallas Clock Chip on the microprocessor board that will most likely need to be replaced.
ROBERTO DIAZ BEDOLLA
lorad m3 ROM ERROR
November 20, 2014 11:31
DOES ANYONE KNOWS WHAT "ROM ERROR" MEANS IN A LORAD MIII SYSTEM?????
THX ALL.
to rate and post a comment
Pete Schliebner
re: lorad m3 ROM ERROR
November 20, 2014 06:30
Check the +5V supply on the motherboard inside the console. I'll check some old service notes and post here if I find anything else.
to rate and post a comment
Larry Underriner
re: lorad m3 ROM ERROR
November 21, 2014 10:09
The 5VDC should be set for 5.2vdc. On the back of the machine is the LVPS (upper left corner). You will see a little access hole in the cover to adjust the pot. The other possible cause could be the Dallas Clock Chip on the microprocessor board that will most likely need to be replaced.
to rate and post a comment
ROBERTO DIAZ BEDOLLA
re: re: lorad m3 ROM ERROR
November 22, 2014 10:41
yes. You are right. Chip dallas does not work more. I just changed it and the system started!
to rate and post a comment
Larry Underriner
re: re: re: lorad m3 ROM ERROR
November 23, 2014 02:07
Glad I could help!