Hello
I have a problem in Lorad M4.
Error message: Time of exposure less than the calculated exposure time available.
I need service manual.
Please help me.
Thanks
Eduardo
BRAZIL
eduardo@elomed.com.br
Hi, this error appear only with HTC, the minimun time for operation of the grid is at 400ms.
Please check it, and if error persist check the HTC Bucky or program in host microprocessor board.
ok Hello
these are the things u check that can cause this error.
1. open bucky to be sure you do not have a defective motor.
if motor does not move by hand repair bucky.
2. check with scope connected to ir micro brd tp7 and ground. see if you get a 9.5 volt reading when exposure is done at manual 30 kvp and 50 mas.
if not tube
3. if machine times out on all AEC auto ( kv, Filter,time)
service manual says IR microprocessor may be bad.
but also defective tube or hv circuit may cause problem.
Note i have also noticed that when i used 2cm bem and did manual exposure it seemed that beam quality droped off at about half of the collimated area away from the chest wall.
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
Eduardo Tognon
Lorad M4 problem, service manual request
February 02, 2011 01:22
Hello
I have a problem in Lorad M4.
Error message: Time of exposure less than the calculated exposure time available.
I need service manual.
Please help me.
Thanks
Eduardo
BRAZIL
eduardo@elomed.com.br
to rate and post a comment
Yuriy Kuznetsov
Re: Lorad M4 problem, service manual request
February 04, 2011 10:26
I have got it, friend :)
to rate and post a comment
cimon ching
Re: Lorad M4 problem, service manual request
February 05, 2011 08:20
check compression height, check you threth hold and check your aec.
to rate and post a comment
Jose Vallejos
Re: Lorad M4 problem, service manual request
February 06, 2011 09:21
Hi, this error appear only with HTC, the minimun time for operation of the grid is at 400ms.
Please check it, and if error persist check the HTC Bucky or program in host microprocessor board.
Good luck.
to rate and post a comment
Anthony Alleyne
re: Lorad M4 problem, service manual request
January 26, 2015 11:07
ok Hello
these are the things u check that can cause this error.
1. open bucky to be sure you do not have a defective motor.
if motor does not move by hand repair bucky.
2. check with scope connected to ir micro brd tp7 and ground. see if you get a 9.5 volt reading when exposure is done at manual 30 kvp and 50 mas.
if not tube
3. if machine times out on all AEC auto ( kv, Filter,time)
service manual says IR microprocessor may be bad.
but also defective tube or hv circuit may cause problem.
Note i have also noticed that when i used 2cm bem and did manual exposure it seemed that beam quality droped off at about half of the collimated area away from the chest wall.
to rate and post a comment
Carlos Antônio França Nascimento
re: re: Lorad M4 problem, service manual request
June 04, 2015 05:34
Hi!
I have a problem in Lorad M4.
Error message:20 and 33.
I need service manual.
Please help me.
Thanks
Carlos
BRAZIL
rxteccomercio@gmail.com
to rate and post a comment