Hi,

I have a problem with auto-collimation in Lorad M-IV Platinium.

The autocollimation is off and the motor blades of collimator are in funcioning for some times or any time. (so crazy). The operator no touch any button and any configuration and collimator functioning alone.

Please any idea ?

1 person liked this
Log inor Register

to rate and post a comment

(27)
(3)
(5)

James McGinty

re: Lorad M-IV Fault Collimation.

May 25, 2012 10:37

Check the compression accessory detection... is the paddle securely mounted?

Log inor Register

to rate and post a comment

(259)
(59)
(3)

Jose Vallejos

re: re: Lorad M-IV Fault Collimation.

May 25, 2012 01:05

yes mi friend...thank's.

Log inor Register

to rate and post a comment

(11)

Cale Clinkenbeard

re: Lorad M-IV Fault Collimation.

May 25, 2012 06:04

Any error codes. Clean/lube the worm gears.