Lorad M-IV system 35 error message

Tim Austin

Lorad M-IV system 35 error message

July 08, 2008 04:55

I have a lorad m-iv that is getting a system 35 error message with very small breasts. I went through the filament calibration and thought that took care of the problem, but a week later the problem resurfaced. Anyone have any ideas?

Thanks

Log inor Register

to rate and post a comment

Tim Austin

Re: Lorad M-IV system 35 error message

July 14, 2008 02:14

I also checked and adjusted the primary filament current because it was a little low. The problem again went away for a week and now has come back again, but has gotten progressively worse to the point that the error shows up anytime you try to take an exposure. Any ideas on the problem and/or possible solution would be greatly appreciated.

Log inor Register

to rate and post a comment

(2)

noel fleck

Re: Lorad M-IV system 35 error message

July 15, 2008 11:43

You probally already know this but in case you don't. Error 35= tube current is below 10% OF EXPECTED VALUE. Check tube current level,calibration,filaments, and protective diodes across filaments.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

We repair MRI Coils, RF amplifiers, Gradient Amplifiers and Injectors.

MIT labs, experts in Multi-Vendor component level repair of: MRI Coils, RF amplifiers, Gradient Amplifiers Contrast Media Injectors. System repairs, sub-assembly repairs, component level repairs, refurbish/calibrate. info@mitlabsusa.com/+1 (305) 470-8013

stats

Upcoming Auction Events

C2 Management
TIMED
Jul 21, 2026 (345 lots)
Surplus Service
TIMED
Jul 22, 2026 (132 lots)
Red Kingdom Wholesale
TIMED
Jul 24, 2026 (115 lots)
ZRG Medical
TIMED
Jul 31, 2026 (218 lots)