Hi guys,

I've got the Error 30 whenever somebody tries to make an exposure. The X ray tube Wwas replaced before and it worked just fine for a few days, but now it won't make an exposure, not in Manual, Auto KV, Auto time and in Auto Filter modes.

Could someone help me here?
Thanks in advance

Log inor Register

to rate and post a comment

(432)
(83)
(3)

juan pablo amado

re: Lorad MIV Error 30

July 20, 2017 10:07

Error 30: error in data from host. Error may indicate software lockup with the host. Origin: image receptor microprocessor

1 person liked this
Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: Lorad MIV Error 30

July 23, 2017 04:45

Thanks Juan Pablo, I will try to test this image receptor on another system.

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Lorad MIV Error 30

July 21, 2017 05:05

Check the image receptor, check the pins on the receptor. Did you try using a different receptor?
Error Message: System Error 30
Source:Image Receptor MP
Description: Host data error
possible Fix: Host software locked-up

1 person liked this
Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: Lorad MIV Error 30

July 23, 2017 04:45

Hi Lenny, I have not tried with another Image receptor, I will though...I had before and error 20 and then error 30.. I checked the tube rotor, and it was going bad already, so I replaced the tube... so when I installed the tube last monday, I did not have any issue... until wednesday that I could not make an exposure.

Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: Lorad MIV Error 30

July 23, 2017 04:45

Hi Lenny, I think I must do calibrations... so you have the service manual with calibrations you can share with me?

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: Lorad MIV Error 30

July 24, 2017 10:42

I need an email address to send you the manuals. Unless the battery dies on the CPU board or you replace a part you should not have to do the calibrations

Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: re: re: Lorad MIV Error 30

July 24, 2017 03:43

francisco.garcia@bioradx.com

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Multi-Vendor Imaging Service Partner

BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729

stats

Upcoming Auction Events

Cbrx2sales
TIMED
Jun 05, 2026 (157 lots)