In auto-time the result is higher mAs. than the one I get regularly for the Phantom that has to be submitted for ACR.
I suspected that I needed to do the GAIN adjustment for the AEC.
I attempted to adjust but even at maximum gain the voltage I get at TP7 is approximately 7.5 volts measured with the oscilloscope as per procedure. (The voltage should be approximately 9.0-9.5 volts)
Is there anyone who can help me with this problem?
Most times I just adjust gain master d
And offset for each kvp station. Just pressing the four switches of right side board.of course it depends on wich chisis I am using. I MEAN kodak fuji ...
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Manuel Brugat
Lorad MIV Platinum AEC Problem
April 06, 2017 10:14
Dear Colleagues:
I am having a problem with a Lorad MIV Platinum.
In auto-time the result is higher mAs. than the one I get regularly for the Phantom that has to be submitted for ACR.
I suspected that I needed to do the GAIN adjustment for the AEC.
I attempted to adjust but even at maximum gain the voltage I get at TP7 is approximately 7.5 volts measured with the oscilloscope as per procedure. (The voltage should be approximately 9.0-9.5 volts)
Is there anyone who can help me with this problem?
Sincerely:
Manuel Brugat/Image Instrumentation, Inc.
to rate and post a comment
alexander grau
re: Lorad MIV Platinum AEC Problem
April 07, 2017 10:20
shoulbe you used caibration AEC with plexiglass, my email is alexgrau1966@gmail.com, speaking spanish
to rate and post a comment
Juan Antonio Ramirez Ramirez
re: Lorad MIV Platinum AEC Problem
April 07, 2017 10:21
Most times I just adjust gain master d
And offset for each kvp station. Just pressing the four switches of right side board.of course it depends on wich chisis I am using. I MEAN kodak fuji ...
to rate and post a comment