Lorad selenia err initializing the DRAPI config files detector missing

Tags: Mammo

(2)

soloana Kautar

Lorad selenia err initializing the DRAPI config files detector missing

July 15, 2019 10:31

Hi
ALL MEMEBRS

I have completed the software installation
And now when I log in.

Application started up and shows

THE DIGITAL DETECTOR IS WARMING UP

on the other window it shows

THERE WAS A PROBLEM INITIALIZING THE DRAPI.
CHECK FOR

1)MISSING CONFIGURATION FILES FOR DETECTOR

2) A CORRUPT GAIN CALIBRATION OR MAP FILE

THE APPLICATION WILL NOW BE RESTARTED!

NOTE : ONE THING MORE I HAVEN'T COPIED ARRAY FILES FROM FTP SERVICE LAPTOP BECAUSE I HAVEN'T any files copied I installed new hard drives due to lost the last system drives

AND IN THIS STATE I JUST INSTALLED SOFTWARE ITS NOT CONNECTED WITH GANTRY.

Waiting for your good reply

Thanks

Log inor Register

to rate and post a comment

(2)
(1)

mike shouppe

re: Lorad selenia err initializing the DRAPI config files detector missing

July 21, 2019 12:59

You will need to install the detector files. Hopefully you have the disk that came with the detector or a backup.