ls qxi

Tags: CT, GE

(62)
(11)

zekeng brice

ls qxi

October 23, 2014 04:16

HI TO ALL please I AM facing a problem on a lighspeed qxi When customer attempted to scan, the rotor accelerated & the scan button activated. When pressed, the xray light illuminated briefly & then two errors were created in the log; 183289 & 245043 (Unexpected loos of high voltage.) and scanner hardware stopped is on the left monitor even resuming don't move anymore

Log inor Register

to rate and post a comment

(11)
(1)

Ed Bican

re: ls qxi

October 27, 2014 10:45

Do you have the JEDI generator or the older version with two inverters. If you have the older version generator check for a blown fuse in either the anode or cathode inverter supplies. It will be a large ceramic type fuse and you will have to take the cover off of the inverter to access it.

Log inor Register

to rate and post a comment

(62)
(11)

zekeng brice

re: re: ls qxi

October 29, 2014 04:32

I have the older version of generator with two inverters thank I will do it and let you know