HI TO ALL please I AM facing a problem on a lighspeed qxi When customer attempted to scan, the rotor accelerated & the scan button activated. When pressed, the xray light illuminated briefly & then two errors were created in the log; 183289 & 245043 (Unexpected loos of high voltage.) and scanner hardware stopped is on the left monitor even resuming don't move anymore
Do you have the JEDI generator or the older version with two inverters. If you have the older version generator check for a blown fuse in either the anode or cathode inverter supplies. It will be a large ceramic type fuse and you will have to take the cover off of the inverter to access it.
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
zekeng brice
ls qxi
October 23, 2014 04:16
HI TO ALL please I AM facing a problem on a lighspeed qxi When customer attempted to scan, the rotor accelerated & the scan button activated. When pressed, the xray light illuminated briefly & then two errors were created in the log; 183289 & 245043 (Unexpected loos of high voltage.) and scanner hardware stopped is on the left monitor even resuming don't move anymore
to rate and post a comment
Ed Bican
re: ls qxi
October 27, 2014 10:45
Do you have the JEDI generator or the older version with two inverters. If you have the older version generator check for a blown fuse in either the anode or cathode inverter supplies. It will be a large ceramic type fuse and you will have to take the cover off of the inverter to access it.
to rate and post a comment
zekeng brice
re: re: ls qxi
October 29, 2014 04:32
I have the older version of generator with two inverters thank I will do it and let you know