hi dear please can anybody help me on ge ls qx/i with the small values less than 100kva and 100ma you can get exposure and exam but above theses values you will get scanner hardware stopped scan kv out of acceptable range and on the err code you will get err 183296, err183183 err183100 I did the gen cal is was hanging but by resuming it was terminated and then the generator characterization n by auto ma with succeful please if someone can bring me a new idea and in this case the tube warmup can't be archieve and the fast cal
You must to do the kv loop test, to determine if you have a bad xray tube or bad HV transformer, fiestas disconnect the HV cables from the tube, and full the receptacles with HV dieléctric Oil, run the test from low HV lets say 80 kv and see if you have any arquing, if you don't have any continúe with 100 kv then 120 and 140 Last if you still have no problem put the HV cables ti the xray tube and run the test again if fácil is present consider a bad tube.
Also you cant run the kv&Ma test in stationary Mode and diferent angles if failure Maybe a bad HV TANK (maybe a bubble inside)...
Lightly used system available direct from the end user customer. MONARCH QUEST
Featuring our most advanced, data-driven navigation software with AI-Powered algorithms & advanced imaging integration with the GE HealthCare OEC 3D C-arm.
zekeng brice
ls qx/i
July 23, 2015 06:14
hi dear please can anybody help me on ge ls qx/i with the small values less than 100kva and 100ma you can get exposure and exam but above theses values you will get scanner hardware stopped scan kv out of acceptable range and on the err code you will get err 183296, err183183 err183100 I did the gen cal is was hanging but by resuming it was terminated and then the generator characterization n by auto ma with succeful please if someone can bring me a new idea and in this case the tube warmup can't be archieve and the fast cal
to rate and post a comment
Lenny Place
re: ls qx/i
July 24, 2015 05:42
you are getting mA and KV errors you sure your tube is not arcing
to rate and post a comment
zekeng brice
re: re: ls qx/i
July 25, 2015 12:54
tanks lenny you may be right I thing something is almost sure or probably the problem should be between the cathode tank and the tube
to rate and post a comment
Ignacio Garcia
re: ls qx/i
July 24, 2015 05:42
You must to do the kv loop test, to determine if you have a bad xray tube or bad HV transformer, fiestas disconnect the HV cables from the tube, and full the receptacles with HV dieléctric Oil, run the test from low HV lets say 80 kv and see if you have any arquing, if you don't have any continúe with 100 kv then 120 and 140 Last if you still have no problem put the HV cables ti the xray tube and run the test again if fácil is present consider a bad tube.
Also you cant run the kv&Ma test in stationary Mode and diferent angles if failure Maybe a bad HV TANK (maybe a bubble inside)...
to rate and post a comment
zekeng brice
re: re: ls qx/i
July 25, 2015 12:54
ok tank you ignacio garcia for your assistance Iwill go back on the site and let you know what happen but Iam suspecting more the cathode tank
to rate and post a comment