Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Lunar DPL

(58)
(6)

Elias Valdes

Lunar DPL

November 30, 2004 03:52

Hello:
Does anybody have a schematics diagram of an OC2 board with an orange dot? We have a problem with a bone densitometer, a Lunar DPL. Both motors, trans. & long, do not work. The motors are not receiving the supply voltage. The problem started after replacing a Bertan Negative High Voltage Power Supply with a Spellman Power Supply, so we have a BertanPositive and a Spellman Negative. We understand this shouldn't cause any problems, is this OK? I woill appreciatte any sugestions. Thank you.

Log inor Register

to rate and post a comment

(139)
(11)
(2)

ALEXANDER PILAFTSIS

Re: Lunar DPL

December 01, 2004 04:22

Dear Mr Valdes,

Replacing HVPSs can not effect at any way OC2 board.
You must follow this steps:
1. Are the Power on led or light on?
2. If yes are the two leds of sbc board illuminated.
3. If yes are the funs of X RAY tube working.
4. If yes then you realy have a problem with oc2 board.
If not there are many-many other resons to have a problem. What is/are the message(s) that you get it will be very helpfull.

Regards

Alexander Pilaftsis