Lunar DPX -NT problem

(28)
(7)
(22)

Semih Gunduz

Lunar DPX -NT problem

July 21, 2009 02:19

Hello everyone,
We have a problem with Lunar DPX NT.
When we are making calibration if the spillliover test avarage is less then 12% in passes the calibration but if the spillliover test avarage is less then 12% it cannot pass the calibration so we need to make mecanical adjustment to get it pass .
Is there any other way to get this done ? becasue mechanical adadjustment takes time!
Thank you for your help.

Log inor Register

to rate and post a comment

(343)
(83)
(13)

Manuel Casanova

Re: Lunar DPX -NT problem

July 21, 2009 05:14

dears:

First adjust the home position for x-ray tube, and then centering the tube between rx and detector.

The problem is the QA block is outside the home position, but only you need calibrate the home position and x-ray tube

Lucky

Log inor Register

to rate and post a comment

(18)
(3)

Mario Mancha

Re: Lunar DPX -NT problem

July 22, 2009 01:59


if you are still having problems after adjust home pos and centering you may have a bad detector

Log inor Register

to rate and post a comment

deepuvijayan84@yahoo.com vijayan

Re: Lunar DPX -NT problem

July 29, 2009 08:22

If you still have problem check the detector first and then check the xray tube also.very first you delete the qa database and done the warmup.if u read the same value for spillover then only check the tube and detector

Log inor Register

to rate and post a comment