JOSE HERRERA

Lunar. DPX.

February 29, 2024 09:55

Dear Sirs.

I have problems in and with the QC calibration of a Lunar bone desitometry machine, DPX, which has this problem:

22293: Low energy reference failure. Low energy reference counts for detector #1 failed. Value:171168. Q.C.
22294: High energy reference failure. High energy reference counts for detector #1 failed. Value:198970 QC

Could you help me?.

Best regards.
Jose

Log inor Register

to rate and post a comment

(37)
(7)
(3)

Albert J. GARRY

re: Lunar. DPX.

March 01, 2024 10:39

Hello Jose,

If you completed a QA, send me a picture of it. Also, do you know how to get into the service mode?


Log inor Register

to rate and post a comment

(12)
(1)

Mohammad Mahmoud

re: Lunar. DPX.

March 01, 2024 10:39

Hi

You need to check the fine alignment between Tube and Detector.

Log inor Register

to rate and post a comment

(28)

ALFATECH

re: Lunar. DPX.

March 04, 2024 10:39

Dear Jose, you have to check the alignment x-ray tube and detector. If it is not correct realign.
If it is correct, then you might have a problem with PMT detector, x-ray tube or shutter.
Most likely with the first one. PMT Detector. In order to be sure you have to check the Persent spillover history.