I have problems in and with the QC calibration of a Lunar bone desitometry machine, DPX, which has this problem:
22293: Low energy reference failure. Low energy reference counts for detector #1 failed. Value:171168. Q.C.
22294: High energy reference failure. High energy reference counts for detector #1 failed. Value:198970 QC
Dear Jose, you have to check the alignment x-ray tube and detector. If it is not correct realign.
If it is correct, then you might have a problem with PMT detector, x-ray tube or shutter.
Most likely with the first one. PMT Detector. In order to be sure you have to check the Persent spillover history.
JOSE HERRERA
Lunar. DPX.
February 29, 2024 09:55
Dear Sirs.
I have problems in and with the QC calibration of a Lunar bone desitometry machine, DPX, which has this problem:
22293: Low energy reference failure. Low energy reference counts for detector #1 failed. Value:171168. Q.C.
22294: High energy reference failure. High energy reference counts for detector #1 failed. Value:198970 QC
Could you help me?.
Best regards.
Jose
to rate and post a comment
Albert J. GARRY
re: Lunar. DPX.
March 01, 2024 10:39
Hello Jose,
If you completed a QA, send me a picture of it. Also, do you know how to get into the service mode?
to rate and post a comment
Mohammad Mahmoud
re: Lunar. DPX.
March 01, 2024 10:39
Hi
You need to check the fine alignment between Tube and Detector.
to rate and post a comment
ALFATECH
re: Lunar. DPX.
March 04, 2024 10:39
Dear Jose, you have to check the alignment x-ray tube and detector. If it is not correct realign.
If it is correct, then you might have a problem with PMT detector, x-ray tube or shutter.
Most likely with the first one. PMT Detector. In order to be sure you have to check the Persent spillover history.