Ampronix, a Top Master Distributor for Sony Medical, provides Sales, Service & Exchanges for Sony Surgical Displays, Printers, & More. Rely on Us for Expert Support Tailored to Your Needs. Email info@ampronix.com or Call 949-273-8000 for Premier Pricing.
Jeremiah Kiblin
Lunar Prodigy - Bone Density
July 07, 2015 09:54
Hello,
Errors:
26738 – Peaking Failure
All Detectors (D1-D16) same error 32776 – Excessive counts during peaking.
22239 – all Detectors failed peaking.
22334 – Spillover failure
22288 – All detectors failed spillover
These are the errors I got trying a QA. I don't have any test tools or manuals. Is the problem with the detector or tubehead ?
If costly repair we might retire unit?
Does any know how much a detector is ?
Does any know how much a tubehead is ?
Can you still get parts for this unit?
Thanks for any help.
Jeremiah
to rate and post a comment
Ben Williams
re: Lunar Prodigy - Bone Density
July 07, 2015 06:22
Hi Jeremiah,
Believe it or not you may not have an issue with the detector or tube at all. Can you send the error log to my email? If so, I will be able to give you a better diagnosis. This string of errors can be caused by a simple issue that is easily overlooked depending on what version of the Prodigy you have.
ben@dexasolutions.com
Thanks,
Ben
(FYI, I am a former Lunar FE and Service Manager)
to rate and post a comment
villa
re: Lunar Prodigy - Bone Density
July 07, 2015 06:22
Jeremiah can you contact on lemedina2000@hotmail.com
Eduardo
to rate and post a comment
Lenny Place
re: Lunar Prodigy - Bone Density
July 07, 2015 07:07
check the aperture assembly, there are rubber o-rings that dry rot and does not allow the aperture to operate correctly
to rate and post a comment
Engr Ali Hussain
re: Lunar Prodigy - Bone Density
July 08, 2015 07:55
Hi jeremiah
I hope you are well.
You can download manual from online GE library.
You can check the X-ray tube by simply placing film on the table and after exposure you will check film on CR DR.
Check the dry sold on boards,
I faced problem of "detector reset eror" whch can be solve by plug out and plug in of detector cable or use blower to clean dust from recoiling slit of tube head.
Parts will be arrange but first make sure which part you need.
You can freely call me on 0092 332 3100896.
Thank you
Regard
Engr. Ali Dawoodi
ali.dawoodi52@gmail.com
to rate and post a comment
Engr Ali Hussain
re: Lunar Prodigy - Bone Density
July 08, 2015 07:55
Hi jeremiah
I hope you are well.
You can download manual from online GE library.
You can check the X-ray tube by simply placing film on the table and after exposure you will check film on CR DR.
Check the dry sold on boards,
I faced problem of "detector reset eror" whch can be solve by plug out and plug in of detector cable or use blower to clean dust from recoiling slit of tube head.
Parts will be arrange but first make sure which part you need.
You can freely call me on 0092 332 3100896.
Thank you
Regard
Engr. Ali Dawoodi
ali.dawoodi52@gmail.com
to rate and post a comment
Jeremiah Kiblin
re: Lunar Prodigy - Bone Density
July 08, 2015 02:41
Thanks for your responses. My Bone Density unit is in the process of getting replaced. I'm trying to find out how soon. If the Radiology Manager wants me to trouble I will contact everyone that responded. Thanks
to rate and post a comment
Jeremiah Kiblin
re: re: Lunar Prodigy - Bone Density
July 09, 2015 09:07
Our new Bone Density is coming in 5-10 days. Not going to troubleshoot this system. If ok, I will keep your contact info for future repairs. Thanks