I HAVE A PROBLEM WITH PRODIGY BONE DENSITOMETER: QA ERROR = RX COMMAND FAILED..... DO YOU HAVE ANY IDEA?? THE PS +28 VDC IS OK..ALL FUSES OK, ..... HOW CAN HELP ME? BYE
go to log files and see the description of error, then go service menu and try to ramp kV, in most of the cases the tube must be replaced, my gues is you have arcing in the tube