Luner Prodigy cannot get past QA mode

Tags: Bone Densitometer

Larry Sprague

Luner Prodigy cannot get past QA mode

October 25, 2013 12:17

System comes up but cannot get past QA mode. It will not let them calibrate. Consequently, then can't use the machine.

Your thoughts.

Thank you

Larry
706-860-0363
imagingresources.xray@gmail.com

Log inor Register

to rate and post a comment

(8)
(7)
(1)

Ricardo Del Olmo

re: Luner Prodigy cannot get past QA mode

October 25, 2013 06:29

What type of error? At what part of the calibration it fails?

Log inor Register

to rate and post a comment

Larry Sprague

re: re: Luner Prodigy cannot get past QA mode

October 29, 2013 08:46

Cannot even start the calibration. We have an idea where the problem might be so we will check it out and see where we go. Thanks for the reply

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Luner Prodigy cannot get past QA mode

October 25, 2013 06:30

what on the QC report is failing?

Log inor Register

to rate and post a comment

Larry Sprague

re: re: Luner Prodigy cannot get past QA mode

October 27, 2013 05:37

They can't get past the start screen of the calibration page.

I'm an x-ray guy, mostly sales on top of that. I did print out all of the error log and it is 70 pages long.

If there is info on the error log that will help let me know and I'll respond.

Log inor Register

to rate and post a comment

Desmond Johnson

re: re: re: Luner Prodigy cannot get past QA mode

October 28, 2013 01:53

As a general rule the first error is the real error and the rest is a dump from the original. Look for an error that has an ARC or HV or RESET in it.

Log inor Register

to rate and post a comment

Larry Sprague

re: re: Luner Prodigy cannot get past QA mode

October 27, 2013 05:38

It will not let them even start a QC.

There is 70 pages of error logs. Don't know where to start.

Log inor Register

to rate and post a comment

Marcelo Antonio Correa

re: Luner Prodigy cannot get past QA mode

October 26, 2013 09:38

Hi, Larry! Look for a tab or option named 'Reset QA Database'. Save prior QA Database and create a brand new one. Sometimes, the QA has some value out of the normal trending that could cause this loss of mean values. Try this, first, because is easy and use to solve. After the first succesful QA, perform at least other 4 QA's, to make sure equipment will have mean values to work. Hope this helps. Take care, God Bless. MAC.

1 person liked this
Log inor Register

to rate and post a comment

stats
DOTmed text ad

PHILIPS Mx-8000 IDT/Brilliance 16 Slice CT Scanner -----Worldwide leader!!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)