Over 150 Missouri Auctions End Today - Bid Now

Machellette Collimaster C

Chris Salberg

Machellette Collimaster C

March 15, 2012 03:26

I need a schematic showing me the resistance value of R1 on collimator lamp timer board. The resistor is toated! Anyone have this? hanks in Advance!

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

Re: Machellette Collimaster C

March 16, 2012 09:11

270 Ohms 1/2 watt

Log inor Register

to rate and post a comment

(265)
(50)
(28)

Wayne Horsman

Re: Machellette Collimaster C

March 16, 2012 09:11

My C-FSR schematic indicates a 270 ohm 1/2 watt resistor.
email me if you want me to send you a scan of the lamp timer pcb schematic.

Log inor Register

to rate and post a comment

Chris Salberg

Re: Machellette Collimaster C

March 17, 2012 09:51

I want to thank all that replied! It is greatly appreciated and the problem has now been resolved although I am very suspicious about the relay on that board. Has anyone found the manufacturer or cross referenced the relay and ordered it? Any further information would be appreciated and thanks to all once more!


Thank you Wayne for the reply. Are you by chance related to Howard Horsman that I worked with at Abbott Northwestern Hospital in Minneapolis? If you get a chance to scan the schematic I would greatly appreciate it! Thanks again!!

Log inor Register

to rate and post a comment

(265)
(50)
(28)

Wayne Horsman

Re: Machellette Collimaster C

March 18, 2012 10:49

I am not related (as far as I know), however the Horsmans (without an "e") are a rare spelling, from Welsh origins. They came down through Nova Scotia long ago, and Minnesota "almost Canada", so who knows?

email me at wayne@columbiaimaging.com so that I send send you the scan of the lamp timer PCB.

Log inor Register

to rate and post a comment

(10)
(1)

antolin cansigno

Re: Machellette Collimaster C

March 16, 2012 09:13

Hi guy, what is the make of the collimator for helping you.

Log inor Register

to rate and post a comment

(6)
(2)

Ronald Lorg

Re: Machellette Collimaster C

March 15, 2012 05:58

R1 is a 270 ohm 1/2 watt resistor. Send me your email for diagrams. r.lorg@verizon.net

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Online auction: Spexis AG Biopharma & Laboratory Equipment

Online auction of modern high-end laboratory and biopharma research equipment from the recent bankruptcy of Spexis AG in Allschwil near Basel, Switzerland. Start bidding today! closes on June 4. info@dome-auctions.com..+31(0)20 808 2880

stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)