After Power ON and during selftesting get message:
"HVC card is not enabled, pres F9 to restart".
If press F9 get message "FATAL FAULT, try turn off and on".
In system setup HVC card enable can not be done (if select nothing is happened).
Before this fault, brest compressor was moved to most up position and now can not put down.
Thanks for any help.
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Milos Krstic
Mammo ELSCINT CH22S problem
February 13, 2010 09:32
After Power ON and during selftesting get message:
"HVC card is not enabled, pres F9 to restart".
If press F9 get message "FATAL FAULT, try turn off and on".
In system setup HVC card enable can not be done (if select nothing is happened).
Before this fault, brest compressor was moved to most up position and now can not put down.
Thanks for any help.
to rate and post a comment
Pi Jaworski
Re: Mammo ELSCINT CH22S problem
February 14, 2010 10:44
Plese check the fuses on the HVC and the presence of the voltages on this board. BTW Elscintec still supports this unit.
to rate and post a comment
Wojciech Selwesiuk
Re: Mammo ELSCINT CH22S problem
February 23, 2010 04:08
Hi,
Field lamp on/off switch is depressed and locked. Check all of them and restart the system.
Good luck