I am writing from Venezuela, south america, I manage the Radiology Department of the Western Health Medical center, Centro medico oriental de salud, we have lost the schematics electronic diagram from the LORAD RT125 PORTATIL xray, so I wonder if you could sale and email us the schematics, our x ray unit have been damage for almost A YEAR and no company in the country have what we need, a engineer needs the schematics diagram to attend repair the unit, waiting your response, thanks in advance
We work with hospitals providing reliable ultrasound systems, probes & parts. By offering sales, service, rentals & repairs, we help our clients access high quality imaging systems while keeping costs low & support simple. Please call 760-834-4777!
DJIELA Flavien
Mammo Lorad M III problem
August 14, 2009 02:57
Dear all,
My mammography system ( lorad M III ) is showing the following error during X ray explosure:
filament power supply
Any help will be appreciated.
to rate and post a comment
preecha chaipet
Re: Mammo Lorad M III problem
August 25, 2009 11:59
hi am preecha ,thailand
filament error , under /over shoot it can not exposure , can you check R42 on TCB read 2.4 amp
you need calibration
Rgds
to rate and post a comment
Jose Garcia
Re: Mammo Lorad M III problem
July 18, 2011 12:02
I have same problem but Dont have service manual..Can You provide this?
Thanks
to rate and post a comment
jgbero
re: Mammo Lorad M III problem
November 15, 2015 09:58
I am writing from Venezuela, south america, I manage the Radiology Department of the Western Health Medical center, Centro medico oriental de salud, we have lost the schematics electronic diagram from the LORAD RT125 PORTATIL xray, so I wonder if you could sale and email us the schematics, our x ray unit have been damage for almost A YEAR and no company in the country have what we need, a engineer needs the schematics diagram to attend repair the unit, waiting your response, thanks in advance