Dear colleagues could you please help me On the unit MammoDiagnost AR Philips. Im having this problem. After the x-ray shot when the compression paddle is going up once or twice a day the unit turn off by itself the main contactor opens, is posible to turn it on again and continue to work. mains Voltage is ok monitored by motor saver at 240 v. there are not errors in the log.
Any idea will be very helpful.
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Allan Morazan
Mammodiagnost AR
August 28, 2016 04:06
Dear colleagues could you please help me On the unit MammoDiagnost AR Philips. Im having this problem. After the x-ray shot when the compression paddle is going up once or twice a day the unit turn off by itself the main contactor opens, is posible to turn it on again and continue to work. mains Voltage is ok monitored by motor saver at 240 v. there are not errors in the log.
Any idea will be very helpful.
thanks.
to rate and post a comment
Lenny Place
re: Mammodiagnost AR
August 30, 2016 09:04
I had a similar issue and found a wire where the insulation had worn away and would intermittently short to ground
to rate and post a comment