marconi eclipse cmedtech rf amp

October 28, 2014 11:42

need help marconi eclipse cmedtech rf amp what is code 89

Log inor Register

to rate and post a comment

(39)
(8)
(9)

Bruce Smith

re: marconi eclipse cmedtech rf amp

October 29, 2014 08:34

89 is an air pressure fault in the PA cavity. Likely either the Blower fan is bad or the pressure sensor itself.

Log inor Register

to rate and post a comment

(268)
(20)
(167)

Marshall Shannon

re: marconi eclipse cmedtech rf amp

October 29, 2014 08:36

FAULT 89
TUBE AIR PRESSURE INSUFFICIENT, IDENTICAL TO ERROR 88

ERROR 88
FAN AIR PRESSURE INSUFFICIENT; IF THE FANS ARE TURNING AND HTE COVERS ARE ON THE TRANSDUCER HAS FAILED, CABLING HAS FAILED OR A/D PROBLEM EXISTS ON PROCESSOR BOARD (UNLIKELY)

LIKELY FAN IS OUT OR COVERS ARE NOT IN PLACE DURING TESTING.

1 person liked this
Log inor Register

to rate and post a comment

(13)
(1)
(6)

Gene Nelson

re: re: marconi eclipse cmedtech rf amp

October 30, 2014 04:37

Thanks for your help

Log inor Register

to rate and post a comment

(39)
(8)
(9)

Bruce Smith

re: marconi eclipse cmedtech rf amp

October 29, 2014 08:36

89 fault is a PA air pressure error. The likely suspects are the Blower and the pressure sensor.

Log inor Register

to rate and post a comment

(9)

mriguy

re: marconi eclipse cmedtech rf amp

October 29, 2014 08:36

code 89:TUBE AIR PRESSURE INSUFFICIENT.
if the fans are turning,and the cover are on,then either the Pressure transduce has failed,there is a cabling problem or an A/D problem exist on the Processor board

Log inor Register

to rate and post a comment

(39)
(8)
(9)

Bruce Smith

re: marconi eclipse cmedtech rf amp

October 29, 2014 04:31

FYI, this has nothing to do with the fans on the front of the amp. If they turn is not relevant to an 89 fault. the blower is an internal part that is not visible with covers on.

Log inor Register

to rate and post a comment

(13)
(1)
(6)

Gene Nelson

re: re: marconi eclipse cmedtech rf amp

October 30, 2014 11:50

Thanks 89 = 88 tube air pressure insufficient squirrel cage
blower was not running found F5 fuse blown all ok thanks for response

Log inor Register

to rate and post a comment

(21)
(3)

Pi Jaworski

re: marconi eclipse cmedtech rf amp

October 30, 2014 11:52

It happened to me several times that removing the sensor and washing it with isopropanol alcohol removed the error. You have to slide out the amp - disconnect all cables at the rear. When you open. the left cover you'll see the small grey plastic piece iwith 2 cables n the botto right.

Log inor Register

to rate and post a comment

(13)
(1)
(6)

Gene Nelson

re: re: marconi eclipse cmedtech rf amp

October 30, 2014 04:37

thanks so much for your reply

Log inor Register

to rate and post a comment

(39)
(8)
(9)

Bruce Smith

re: re: marconi eclipse cmedtech rf amp

October 31, 2014 08:57

This is not a recommended practice. this typically shorts the sensor and overrides the air air flow protection that was intended to keep your RF tubes from burning up. The sensors are inexpensive, we will sell you a new one so you don't risk tube damage.

Log inor Register

to rate and post a comment

(21)
(3)

Pi Jaworski

re: re: re: marconi eclipse cmedtech rf amp

October 31, 2014 05:12

I did not suggest to remove the sensor (I am pretty sure, the amp checks both OFF and ON states) but to clean the sensor with IPA and put it back. It works.

Log inor Register

to rate and post a comment

(39)
(8)
(9)

Bruce Smith

re: re: re: re: marconi eclipse cmedtech rf amp

November 01, 2014 02:07

The amp only know if that is open or short, short is the "Air is fine" condition. IF this sensor shorts there is no protection for the air pressure and the amp will not know.

Log inor Register

to rate and post a comment