It means "Anode over voltage". It seems like a HV spike. First thing you can do is HV connectors cleaning and greasing with silicone oil. Check it before for damage. Write if you need more help.
We are the best choice for all your Endoscopy needs. We have Certified Pre-Owned refurbished and used Endoscopes, and all other related components and
accessories. We are committed to your absolute satisfaction. (513) 469-7267
Dionicio Duran
Marconi SeleCT SP Error code
October 08, 2009 04:26
Please any one help to decode FOLLOWING error code on a Marconi SeleCT SP CT Scanner.
error CODE: ER80, 3C92, 2, 78, 64,, A,6,5,9FE8, B380
Thanks
Dion Duran
email: dduran@medintek.com
to rate and post a comment
Ronnie Aguiar
Re: Marconi SeleCT SP Error code
October 09, 2009 05:23
the error happens during the shooting? what values of voltage, current and time?
to rate and post a comment
Wojciech Selwesiuk
Re: Marconi SeleCT SP Error code
October 09, 2009 05:23
It means "Anode over voltage". It seems like a HV spike. First thing you can do is HV connectors cleaning and greasing with silicone oil. Check it before for damage. Write if you need more help.