Over 700 Total Lots Up For Auction at Three Locations - TX 07/30, CA 07/31, MO 08/04

MEDISON 9900 ERROR CODE 31

(147)
(7)
(6)

rossana diaz

MEDISON 9900 ERROR CODE 31

September 12, 2012 11:41

Hello i have a MEDISON 9900, dont have the 4d convex probe so i use another probe from an XQ the model is 3D4-7EK, the system admited the probe on 2d with no issues but when i tried the 4D i got this mesagge in a window

"PROBE DOESNT STAY AT NULL POSITION"

and then

"LIVE 3D SCAN FAILED TCLAYER IS NOT RESPONDING"



then ERROR CODE 31 TURN OFF THE SYSTEM



I s this because is not a coMpatible probe?

PLEASE ANY HELP WILL BE APRECCIATE

Log inor Register

to rate and post a comment

(39)
(7)
(1)

Raymundo Miranda

re: MEDISON 9900 ERROR CODE 31

September 25, 2012 01:21

Yes the error is related to the vm board, do not decode the signal video from 3D/4D proses