Dear Colleagues,
I have a Medison Sonoace 8000 Live that stuck near to the end of loading bar and cannot enter on scanning screen. A few moment later system gives an error message: SYSTEM ERROR Please powerdown system and reboot, error code:64. Is this any problem related to software issues or electronic faulty on system boards? Any kind of help shall be much appreciated.
Best regards!
Le Can
Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net
Le Can
Medison SA8000 Error code 64
March 03, 2014 08:56
Dear Colleagues,
I have a Medison Sonoace 8000 Live that stuck near to the end of loading bar and cannot enter on scanning screen. A few moment later system gives an error message: SYSTEM ERROR Please powerdown system and reboot, error code:64. Is this any problem related to software issues or electronic faulty on system boards? Any kind of help shall be much appreciated.
Best regards!
Le Can
to rate and post a comment
Geri
re: Medison SA8000 Error code 64
March 04, 2014 08:04
Check the power supply. Probably you need only to clean the power supply from the dust? Then you have to check the DSC Board.
Regards
to rate and post a comment
Le Can
re: re: Medison SA8000 Error code 64
March 04, 2014 10:41
Thank you!