Laura Carolina Gonzales Calderon

MEDISON SA8000 ERROR CODE 66

May 06, 2016 03:15

Regards, i need your help, medison sonoace 8000 appear in screen error code 66, This message to which it refers, I estimate very much his help, thaks in advance ing.lauragc@hotmail.com

Log inor Register

to rate and post a comment

Ricardo A. Sanchez

re: MEDISON SA8000 ERROR CODE 66

May 08, 2016 06:04

Laura, Error Code# VM_DSP_ACK_ERROR 66 VM, check the Video Manager board

Log inor Register

to rate and post a comment

(19)
(9)

Hommy Figueroa

re: MEDISON SA8000 ERROR CODE 66

May 08, 2016 06:04

Hi Laura:

Error 66 in a Medison SA8000 refer to a failure in the Video Manager Board. Please perform a swapping test with other VM Board.

Good luck .!

Hommy Figueroa
Alpha Biomedical & Diagnostic Corp.

Log inor Register

to rate and post a comment

(195)
(47)
(4)

EROTOKRITOS SKENTOS

re: MEDISON SA8000 ERROR CODE 66

May 08, 2016 06:06

Laura the first think you have to do is to remove all the boards and re connect them.8000 is very sensitive and most problems i have is just bad connection.If that does not help i will give you further instructions.

Log inor Register

to rate and post a comment

(332)
(69)

Gabriel Cool

re: MEDISON SA8000 ERROR CODE 66

May 08, 2016 06:06

This error comes when you start the machine?

Log inor Register

to rate and post a comment

stats
DOTmed text ad

We provide sales, service & repair for portable ultrasound systems & probes

If you have a broken Sonosite, Mindray, GE, Philips, Hitachi, or Aloka system or probe contact us today. Free ultrasound & probe EVALS with 1 Year Warranty on all repairs. We repair all orders in 1-5 days. Call 844-644-5440 Email: Sales@Lenard.Healthcare

DOTmed Certified stats
advertisement

Upcoming Auction Events

Cbrx2sales
TIMED
Jun 05, 2026 (157 lots)