Medison SA-6000 CMT Software Problem

Posted by an Anonymous User

Medison SA-6000 CMT Software Problem

April 01, 2013 05:35

we have two Medison Sonoace 6000cmt ultrasound systems. when we want to fill a new patient record or delete old patient data, system frozen. no reactions and no workflow on the system. there is no problem ultrasound image or imaging. we try to work with another system software (ghosted hdd). but color mode, harmonıc, pwd and 3D options don't work. how can we reinstall the software or change the HDD. thanks for now.

Log inor Register

to rate and post a comment

(39)
(1)

Anthony Alleyne

re: Medison SA-6000 CMT Software Problem

April 02, 2013 11:11

these systems usually have hard drives that have multiple partitions. one section for the OS and another for the patient image files and one for the temp files. Have your I.T guy read the drive without booting it and you will see that the patient image drive is full. there is usually a service utility to maintain the drive, this often requires password access.

in the interim you can try deleting some of the image files to free up space but be very carefull to image the drive before any changes.

Log inor Register

to rate and post a comment

(26)
(4)

Kadir AKMAN

re: re: Medison SA-6000 CMT Software Problem

April 03, 2013 07:44

I have ghosted image of the original drive. there is no problem, i can get the image. but when i press ID and fill protocol number or name, system frozen about 10 minute. or when i want to delete patients, it's tehe same of before. system frozen and no function and no reactions from keyboard or others.

Log inor Register

to rate and post a comment

(111)
(26)

German Flores

re: re: re: Medison SA-6000 CMT Software Problem

April 03, 2013 01:30

Hi Kadir, you must be load a new ghost to HDD! Besides, you must to load new passwords (COLOR DOPPLER, IMAGE FILE and HARMONIC).
If you can we can help you (almmed@gmail.com).
Regards,
German.

1 person liked this
Log inor Register

to rate and post a comment

(26)
(4)

Kadir AKMAN

re: re: re: re: Medison SA-6000 CMT Software Problem

April 04, 2013 08:11

i have another sytem software, it's working. how can i applicate this software to crashed system. thanks.

Log inor Register

to rate and post a comment

(111)
(26)

German Flores

re: re: re: re: re: Medison SA-6000 CMT Software Problem

April 04, 2013 10:46

Is the same SW version? If it so, you can apply the ghost, then you will need the all paswords to install the optionals. You must input in Service Mode. If you can we can help you (almmed@gmail.com).
Regards,
German.

1 person liked this
Log inor Register

to rate and post a comment

Servicio Tecnico

re: re: re: re: Medison SA-6000 CMT Software Problem

December 11, 2013 06:08

Dear German, please can you help me, we have a Medison SA-6000 with HDD damged, we need to load a new ghost in new HDD to start the machine, but we don't have the software, can you send me the software to do the ghost? (falvarado@diopsa.co.cr). Regards, Francisco Alvarado

Log inor Register

to rate and post a comment

(1)

Carlos Aponte

re: Medison SA-6000 CMT Software Problem

June 01, 2018 09:28

Please someone can help me? I need the Software for a ultrasound Medison SonoAce 6000C, my email is: caponte.aponte@gmail.com
thanks for help