Medrad Spectris Failure - Error Information

April 13, 2016 06:04

SYSTEM DIAGNOSTICS FAILURE - HALTING SYSTEM ERROR DATA: 391,84,6,3006

Deatails : Syringe B failure, working noise no movment ...

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Medrad Spectris Failure - Error Information

April 15, 2016 07:46

have you checked the flex shafts?

Log inor Register

to rate and post a comment

(71)
(13)

Ramil Mirshanov

re: Medrad Spectris Failure - Error Information

April 15, 2016 07:48

This message is displayed if the system does not pass self diagnostic tests.
HCU software has terminated an injection due to an under volume condition in syringe B. Need -
• Remove and re-apply power to the DCU.
• Re-calibrate the injector head potentiometer constants.
• If the condition continues, replace the HCU Power card.
• If the condition continues, replace and calibrate the injector head potentiometers.