This message is displayed if the system does not pass self diagnostic tests.
HCU software has terminated an injection due to an under volume condition in syringe B. Need -
• Remove and re-apply power to the DCU.
• Re-calibrate the injector head potentiometer constants.
• If the condition continues, replace the HCU Power card.
• If the condition continues, replace and calibrate the injector head potentiometers.
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
Nikolaos Moschopoulos
Medrad Spectris Failure - Error Information
April 13, 2016 06:04
SYSTEM DIAGNOSTICS FAILURE - HALTING SYSTEM ERROR DATA: 391,84,6,3006
Deatails : Syringe B failure, working noise no movment ...
to rate and post a comment
Lenny Place
re: Medrad Spectris Failure - Error Information
April 15, 2016 07:46
have you checked the flex shafts?
to rate and post a comment
Ramil Mirshanov
re: Medrad Spectris Failure - Error Information
April 15, 2016 07:48
This message is displayed if the system does not pass self diagnostic tests.
HCU software has terminated an injection due to an under volume condition in syringe B. Need -
• Remove and re-apply power to the DCU.
• Re-calibrate the injector head potentiometer constants.
• If the condition continues, replace the HCU Power card.
• If the condition continues, replace and calibrate the injector head potentiometers.
to rate and post a comment