Hi I thankful to your positive response. Already I check it but I check it again on your instruction. can you guide me How can I verify flash lamp, high voltage supply and which are u guide me.
Hi, for the simmer it is possible to see if the flash lampe light up at the machine is in the ready mode.
For the high voltage, it is possible for you to go in service mode and apply the value for the HV and mesure at the capacity with the high voltage probe and the multimiter if this value is the same?
Are you have the service manuale for this machine?
Hi, I 100% agree with your suggestions. Because you guide me absolutely right way which will be possible you. But my bad luck, I don't have service manual and not available service, mode in this unit.
We are the best choice for all your Endoscopy needs. We have Certified Pre-Owned refurbished and used Endoscopes, and all other related components and
accessories. We are committed to your absolute satisfaction. (513) 469-7267
Mega Pulse 30 laser
December 18, 2017 03:20
Hi Dear
Any one help me about Richard Wolf Mega Pulse 30 laser error.
Simmering
Low water
Plz guide me
srizhaider@gmail.com
to rate and post a comment
salim ouchenir
re: Mega Pulse 30 laser
December 19, 2017 03:57
Hi, I work on candela laser, simmering is the flash lampe of the head laser not work.
The laser nead the water for heat the head to 64 °c, and the machine monitor the température, the level and the flow off the water chek if not the leak or the heater not work.
to rate and post a comment
Riz Haider
re:re: Mega Pulse 30 laser
December 21, 2017 09:18
Hi
I thankful to your positive response.
Already I check it but I check it again
on your instruction. can you guide me
How can I verify flash lamp,
high voltage supply and which are u guide me.
to rate and post a comment
salim ouchenir
re: re:re: Mega Pulse 30 laser
December 21, 2017 04:00
Hi, for the simmer it is possible to see if the flash lampe light up at the machine is in the ready mode.
For the high voltage, it is possible for you to go in service mode and apply the value for the HV and mesure at the capacity with the high voltage probe and the multimiter if this value is the same?
Are you have the service manuale for this machine?
to rate and post a comment
Riz Haider
re:re: re:re: Mega Pulse 30 laser
December 23, 2017 09:50
Hi, I 100% agree with your suggestions. Because you guide me absolutely right way which will be possible you. But my bad luck, I don't have service manual and not available service, mode in this unit.
to rate and post a comment
GV-Felipe Gonzalez
re: Mega Pulse 30 laser
June 02, 2021 02:12
Good morning, will anyone have the Richard Wolf megapulse 30 service manual? Greetings