Mercury 332 error 950

Tags: Radiology - R/F, VILLA

(20)
(2)

Adrian Mihaita

Mercury 332 error 950

October 04, 2022 10:01

Hello,
I'm in trouble with this error, it arises when the table is in a vertical position, and try to move the column to the head end.
For a second the LED displays go off and after that, the error 950 message is displayed.
To the foot end, there is no problem with the motion, or if the table is in a horizontal position.
I'm a little bit confused because from time to time I did the movement without any error.
Did anyone encounter a similar problem?
Any advice will be appreciated!

Log inor Register

to rate and post a comment

(17)
(1)

Peter Yli

re: Mercury 332 error 950

October 06, 2022 09:41

Dear Adrian, since I got the related Serv.Manual, Err.950 is due to an interruption of serial-data-tranfer to the main processor ... it looks like an almost broken wire (false contact) .. or a position potmeter ruined ... Just for info I could arrange the subject manual for you ... but it is written in Italian language ... please confirm writing to my pers. address : yliroma (at) gmal (dot) com. Regards : Peter

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)