MIDMARK M 11 ERROR CODE 005

(37)
(3)
(27)

Francisco Jaramillo

MIDMARK M 11 ERROR CODE 005

May 28, 2013 06:40

some one has experience with this. I have service manual, i made calibration process doesnt work

Log inor Register

to rate and post a comment

(7)

Richard Fernandez

re: MIDMARK M 11 ERROR CODE 005

May 29, 2013 09:41

if you have made the calibration procedure and it has not remedied the issue try cleaning the temperature probe with a mild abrasive pad.
Also check or replace the bellows. if all this fails replace the temp probe.
it would help alot if you can tell me the conditions at the time of failure i.e. chamber pressure. and temp. and condition of door gaskets (door and dam)
Regards,

Rich

Log inor Register

to rate and post a comment

(37)
(3)
(27)

Francisco Jaramillo

re: re: MIDMARK M 11 ERROR CODE 005

May 30, 2013 09:47

Hi Richard i really apreciated your help. I put a new below, i clean the temp probe, i made the calibration procedure, the problem is just when the pressure is around 28 psi ( 198 kpa) and temp is around 120 then show ERROR CODE 5, door gaskets are okey. See like the temp probe, what do you think?

Log inor Register

to rate and post a comment

(7)

Richard Fernandez

re: re: re: MIDMARK M 11 ERROR CODE 005

May 31, 2013 11:13

Ok, I am assuming the 120 Degree is in Celsius.and we know that there are no pressure leaks, start a sterilization cycle, With an multimeter monitor the resistance on the temp probe as it heats up make sure that the it does not go open at any time during the cycle.
Also i am starting to suspect a faulty thermostat at the bottom of the chamber test it in the same fashion .

give me a call at 305 978-8031 perhaps i can be of more help over the phone...if I don't pick up the phone leave me a call back number.

Log inor Register

to rate and post a comment

(37)
(3)
(27)

Francisco Jaramillo

re: re: re: re: MIDMARK M 11 ERROR CODE 005

May 31, 2013 03:22

Hi Richard, for sure the temp is in Celsius, for sure not pressure leaks, i will check the sensor doesn´t open. Very good point of the termostat i will double check, i really appreciated your time and help, my number is +52 5550681413, thanks a lot

Log inor Register

to rate and post a comment

(8)

Marcio Braga

re: re: re: re: re: MIDMARK M 11 ERROR CODE 005

May 31, 2013 06:15

hi, friends.If you calibrated and is giving this error, probably the valve BELLOWS is dirty or clogged.steam is coming out of the trap 90C?

Log inor Register

to rate and post a comment

(37)
(3)
(27)

Francisco Jaramillo

re: re: re: re: re: re: MIDMARK M 11 ERROR CODE 005

June 03, 2013 01:52

MArcio, thanks a lot for your answer, well my friend
i put a new bellow, i already order a new temp sensor and a new termostat, just when i will get it i put and then i tell you if works
thanks
francisco

Log inor Register

to rate and post a comment