MIII Control error

(99)
(11)
(6)

Yojin Ramones

MIII Control error

May 02, 2007 01:15

Hi people I have problem with MIII 1995, When I will shoot appear control error. after that I go to service calibration and start and appear KV sense control error. the KV card dont have any led on . Please let me know what is the problem??

regards

yojin Ramones

Log inor Register

to rate and post a comment

(93)
(7)

Edward Yan

Re: MIII Control error

May 02, 2007 02:37

Control Error means the KV sense exceeds +/-12.5% of setting during exposure.
There is no KV card. It is the High Voltage Generator Control Board. Besides KVP off calibration can cause Control error, the mA overshoot can also cause Control error. To calibrate mA, connect a volt meter across R42 on the Tube Control Board and Connect a storage oscilloscope to the ungrounded end of the 10 ohm calibration resistor.... If you need additional help, contact me at edwardhyan@yahoo.com e-mail address.

Edward Yan, B.S.E.E.

Log inor Register

to rate and post a comment

(99)
(11)
(6)

Yojin Ramones

Re: MIII Control error

July 24, 2008 06:33

sorry for the answer delay but thanks a lot for the information

regards

yojin

Log inor Register

to rate and post a comment

(16)

Anthony Corpstein

Re: MIII Control error

May 02, 2007 03:14

Check small fuse on side of HV module under clear plasic cover - I think it is 1/10 amp it has a spring and resistor inside.