Hi people I have problem with MIII 1995, When I will shoot appear control error. after that I go to service calibration and start and appear KV sense control error. the KV card dont have any led on . Please let me know what is the problem??
Control Error means the KV sense exceeds +/-12.5% of setting during exposure.
There is no KV card. It is the High Voltage Generator Control Board. Besides KVP off calibration can cause Control error, the mA overshoot can also cause Control error. To calibrate mA, connect a volt meter across R42 on the Tube Control Board and Connect a storage oscilloscope to the ungrounded end of the 10 ohm calibration resistor.... If you need additional help, contact me at edwardhyan@yahoo.com e-mail address.
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Yojin Ramones
MIII Control error
May 02, 2007 01:15
Hi people I have problem with MIII 1995, When I will shoot appear control error. after that I go to service calibration and start and appear KV sense control error. the KV card dont have any led on . Please let me know what is the problem??
regards
yojin Ramones
to rate and post a comment
Edward Yan
Re: MIII Control error
May 02, 2007 02:37
Control Error means the KV sense exceeds +/-12.5% of setting during exposure.
There is no KV card. It is the High Voltage Generator Control Board. Besides KVP off calibration can cause Control error, the mA overshoot can also cause Control error. To calibrate mA, connect a volt meter across R42 on the Tube Control Board and Connect a storage oscilloscope to the ungrounded end of the 10 ohm calibration resistor.... If you need additional help, contact me at edwardhyan@yahoo.com e-mail address.
Edward Yan, B.S.E.E.
to rate and post a comment
Yojin Ramones
Re: MIII Control error
July 24, 2008 06:33
sorry for the answer delay but thanks a lot for the information
regards
yojin
to rate and post a comment
Anthony Corpstein
Re: MIII Control error
May 02, 2007 03:14
Check small fuse on side of HV module under clear plasic cover - I think it is 1/10 amp it has a spring and resistor inside.
to rate and post a comment