If you have a broken Sonosite, Mindray, GE, Philips, Hitachi, or Aloka system or probe contact us today. Free ultrasound & probe EVALS with 1 Year Warranty on all repairs. We repair all orders in 1-5 days. Call 844-644-5440 Email: Sales@Lenard.Healthcare
Danilo Abdo
Millennium MPR Detector Calibration
December 20, 2016 02:20
Hello All!
I get here a instalation from ËcoldË, it is I mean, from zero.. Cristal with no tubes, nothing.
So, I install all tubes ( 48 tubes ) and Matrix, IN board, All cables, all necessary, whith a GX200. ( software 4.5 )
So, what I can`t find a solution, is because many tubes ( some times 21 tubes ) becomes out of range, very high. All this tubes already are in low on adjustment.
Can someone give me a help on this?
to rate and post a comment
Lenny Place
re: Millennium MPR Detector Calibration
December 21, 2016 10:48
Have not worked that much on the GE line but on Siemens and Adac I would be verifying my HV power supply and lv supplies. Run the proactive PMT diagnostics and check results
to rate and post a comment
Danilo Abdo
re: re: Millennium MPR Detector Calibration
December 28, 2016 09:01
Thanks Lenny, but the power supply is ok.. I change it 2 times to make sure.. and I can run proactive PMT is not have any pre map on Genie Systems....
to rate and post a comment
BENITO MONTOYA VARELA
re: Millennium MPR Detector Calibration
December 27, 2016 10:15
Danilo , on the case of PM´t some times is written the H.V. PMT check this and si if it is the correct, if you have the same ,please disconnect the 1/2 of the PMTubes and check the others, ,and do the same with the other ,a badPMT can cause your problem , . Also check the type of PMT using in this systems.Suerte
to rate and post a comment
Danilo Abdo
re: re: Millennium MPR Detector Calibration
December 28, 2016 09:01
Hello Benito
What you mean with disconnect 1/2 tubes? PMTUNE will works on half tubes on detector? The PH from tubes is about minimum 160-200mv , and one or 2 tubes on 230~ mv.
Can you help me on it?
thanks!