I have a MPG 65 generator that:
on start up appear error 501. Reset and works
- it works in fluoroscopy
- in radiography mode, on preparation, error 564.
Please, can you help me or do you have a error list. In my manula there are not error code list.
The error 501 is that the cpu battery is bad and the error 564 is absent current of the filament, change the cpu battery and then check the other error my email is mcxrayca@Hotmail.com by any other information
Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.
SBService
MPG 65 GE generator error 564
May 05, 2015 04:22
I have a MPG 65 generator that:
on start up appear error 501. Reset and works
- it works in fluoroscopy
- in radiography mode, on preparation, error 564.
Please, can you help me or do you have a error list. In my manula there are not error code list.
to rate and post a comment
DMEDS
re: MPG 65 GE generator error 564
May 05, 2015 05:43
i have the manuals if you need them contact me ,
magmesme@gmail.com
Regards
to rate and post a comment
Aaron Bassham
re: MPG 65 GE generator error 564
May 06, 2015 08:46
You can find GE Service Manuals for most systems here http://www3.gehealthcare.com/en/support/support_documentation_library
to rate and post a comment
SBService
re: re: MPG 65 GE generator error 564
May 07, 2015 10:43
Thank you, I know, but there are not MPG manuals on GE library.
to rate and post a comment
LUIS RIVAS
re: MPG 65 GE generator error 564
May 06, 2015 08:47
Hello Sbservice,
The error 501 is that the cpu battery is bad and the error 564 is absent current of the filament, change the cpu battery and then check the other error my email is mcxrayca@Hotmail.com by any other information
to rate and post a comment
SBService
re: re: MPG 65 GE generator error 564
May 07, 2015 10:43
Thank you. I will change the CPU battery and after I will contact you.
to rate and post a comment
LUIS RIVAS
re: re: re: MPG 65 GE generator error 564
May 07, 2015 02:29
ok that God blessed to you and solve the fail
to rate and post a comment
Muhammad Muhammad
re: MPG 65 GE generator error 564
August 20, 2017 12:12
i would be very thankful, if someone could send me service manual of MPG 65 xray generator ge healthcare.
thanks in advance
to rate and post a comment