Over 150 Missouri Auctions End Today - Bid Now

MPH M3710 rotor error

Mohamed Ibrahim

MPH M3710 rotor error

March 21, 2010 01:59

hello friends,
I am facing this error in MPH generator after pressing exposure switch , the error is : M3710 which says overcurrent in rotor CTRL during high speed acceleration.
Rotor test gives the following err:
err 239 U010 AD 0000 ED 0000 AA0000
any help is so appreciated.
thnx in advance

Log inor Register

to rate and post a comment

(373)
(66)
(15)

yuan bo

Re: MPH M3710 rotor error

March 22, 2010 09:18

first of all check the rotor board, if it is ok, i think you need to replace tube.

Log inor Register

to rate and post a comment

Mohamed Ibrahim

Re: MPH M3710 rotor error

March 22, 2010 11:10

well, rotor board is ok, and tube is new.

Log inor Register

to rate and post a comment

(244)
(19)

Staate Hayward

Re: MPH M3710 rotor error

March 23, 2010 12:15

http://apps.gehealthcare.com/servlet/ClientServlet

Error 710
with the 3 in front, it may think it's supposed to have a third tube, hence the error.