Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

MRI Contour 3 Table problem

Tags: MRI

(128)
(10)
(5)

Manuel Esquer

MRI Contour 3 Table problem

January 25, 2013 11:41

Table does not go up.

Status on console monitor displays:
"Landmark is lost due to SRP Reset
Move table all the way out, then establish landmark"

Viewlog:
Error 2240106
"The cradle must be at outlimit when you want to up/down the table
Please move the cradle manually, and make sure the outlimit LED is turned on"

Table is outlimit, outlimit sensor works and led on display is turnes on i can see on SRP assy INTR LED turns on when outlimit optical sensor is actuated, and I can check that RL1 is energized when I press UP button on panel. But table does not comes up.
Any idea about solution?