ERROR 2031/243
First,
I did the recovery image on a new hard disk following the service manual and now I have an error of communication between FL-C and FD.
Second,
When I do the test of communication(ping) from the FL-C to the IOBOX (192.168.2.21) and detector (192.168.2.35) it doesnt work.
Third,
Following the service manual it shows me about the VIVA tool for the verification of the communication but in the VIVA tool I can't see to select the serial number of the FD.
So I can't run the dark image test. I suspect that i should install the software paxsanl06 for Varian.
I also need to know if this software is generic or is unique for each detector
We have installed few Siemens Digital Radiography X-ray system. I can try to support. Send me the details to shiju@tecsila.com or Whatsapp at +91 9444026607
Altima will inspire the trust, loyalty and commitment of our customers through positive relationships and collaboration, top quality equipment services, and a clearly superior value proposition. Call Today (469) 644-7063
MIGUEL ANGEL TORRES MARTINEZ
MULTIX FUSION
July 20, 2017 10:03
ERROR 2031/243
First,
I did the recovery image on a new hard disk following the service manual and now I have an error of communication between FL-C and FD.
Second,
When I do the test of communication(ping) from the FL-C to the IOBOX (192.168.2.21) and detector (192.168.2.35) it doesnt work.
Third,
Following the service manual it shows me about the VIVA tool for the verification of the communication but in the VIVA tool I can't see to select the serial number of the FD.
So I can't run the dark image test. I suspect that i should install the software paxsanl06 for Varian.
I also need to know if this software is generic or is unique for each detector
to rate and post a comment
Shiju Karakkoottil
re: MULTIX FUSION
July 21, 2017 10:25
We have installed few Siemens Digital Radiography X-ray system. I can try to support. Send me the details to shiju@tecsila.com or Whatsapp at +91 9444026607