Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

multix pro error code : Err 0002

Tags: Radiology - R/F

Posted by an Anonymous User

multix pro error code : Err 0002

August 02, 2017 07:58

greetings
i have problem with bucky x_ray :
multix pro
siemens
error code : Err 0002
any support please

Log inor Register

to rate and post a comment

(84)
(16)

fortronik

re: multix pro error code : Err 0002

August 02, 2017 11:37

Err002: D300 MCB Error 0
Description:
• SW error or D300 MCB defective.
Possible Causes of Error:
• D300 MCB3 defective
• FW on MCB3 is not compatible with the FW in the Touch Panel.
Corrective Measures:
• Check the D300 MCB FW and Touch Panel per the Compatibility List.
• Download the FW again per the Compatibility List to the D300 MCB.
• Replace D300 MCB3

Log inor Register

to rate and post a comment

yassin kerdaoui

re: re: multix pro error code : Err 0002

August 03, 2017 01:47

thanks sir now the consol show error 0032 and some time 0001.how can I download MCB 300 FW Can you help me

Log inor Register

to rate and post a comment

(21)
(2)

P402

re: re: re: multix pro error code : Err 0002

August 04, 2017 08:53

Replace D500

Log inor Register

to rate and post a comment

yassin kerdaoui

re: re: re: re: multix pro error code : Err 0002

August 05, 2017 04:24

is there some one can help me . Err0032 .Err0031????

Log inor Register

to rate and post a comment

yassin kerdaoui

re: re: re: re: multix pro error code : Err 0002

August 05, 2017 04:25

error 0031 and 0032 any body can help me

Log inor Register

to rate and post a comment

(84)
(16)

fortronik

re: re: re: re: re: multix pro error code : Err 0002

August 06, 2017 04:48

Maybe you have a broken 5.1 (V) power supply or too much resistance on line 5.1 (V) and it does not come to PCB D 500.
There may also be a bad contact of the fuse on the D501 for 5.1 (V)

Log inor Register

to rate and post a comment

(133)
(35)
(1)

Arnie Mataro

re: multix pro error code : Err 0002

August 07, 2017 10:22

External RAM defective. ==> Replace D50.

Log inor Register